Hi,
The statement is in brackets because it's a nomenclature translation, but I'll see if
I can emphasise the non literal nature a bit more.
> My understanding is that the i1Pro1 has a tungsten filament incandescent light source (so
> basically Illuminant A - which would therefore have little violet or blue energy and, I would
> have thought, very little ultraviolet). Given what you say above, I guess that the diffraction
> grating efficiency covers a wider range of wavelengths so that it is at least able to measure
> what little light is reflected back at those wavelengths.
The i1pro2 seems very similar to the i1pro in fundamentals. The illuminant and optics
appear unchanged, apart from the addition of two supplemental UV LEDS. While incandescent
illumination has relatively low UV output, this is magnified in an instrument setting
because the reflected light level is divided by the illuminant level. The result is
that a quite noticeable level of fluorescence can be excited.
> I'm also aware that the i1Pro1 (and the ColorMunki spectro, and likely other spectros as well)
> give best results at higher light intensities and worst at lower ones; so they have problems
> measuring dark colors.
Every instrument does. The i1d3 is significantly better because it collects light
from a much larger area than usual with these low end instruments.
> If that is all correct, then it would seem that the accuracy and precision in the near-uv,
> violet and blue region would be better, on the i1pro1, for emissive sources than for reflective
> ones.
Perhaps, but the diffraction grating and possibly fibre optics seems to limit it.
> I gather that the i1Pro2 has a dual illuminant - a tungsten filament incandescent plus a UV LED
> and that its 'dual scan' mode takes two sets of measurements, one with each illuminant, then
> computes what the results would have been if the illuminant had been D50.
Yes. It's a similar idea to Argyll's FWA compensation, and using a UV
source that you can switch on and off was something I suggested in
the related paper. Nice to see that X-Rite have taken the suggestion up :-)
My initial experiments were of a very similar nature - comparing UV cut
to UV included, and 'A' to simulated D65 illiuminant using the Spectrolino.
> I came across an interesting industrial spectro where the spectro itself and the illuminant are
> entirely different devices, and both connect to the reading head via optical fibres. One of
> their illuminators used four different LEDs (a warm white, cool white, cyan, and blue/uv) to
> approximate a relatively even spectrum.
There are now some viewing booth lights that use a whole set of LEDs to allow you
to dial the spectrum up. I think X-Rite may be doing a similar thing in one
of their high end spectros.
Graeme Gill.