We have followed your suggestion to run a MiSeq test run prior to NovaSeq and are going through your list of QC metrics…
I am now figuring what, if anything, might improve performance on the later platform.
Two MiSeq runs:
Run 1
PhiX 1%
Clustered but PhiX did not workIllumina suggested overloading of library (displaced PhiX)
Without a clear positive control, we repeated the MiSeq with altered amounts of library and PhiX.
Run 2
16 Zeiss LCM caps (one no-template) with non-precious sample tissues (nervous tissue, liver)
Pooled cDNA libraries (AMPure SPRI purification) with unique i5 indexing approach.

Library type: SMART-3Seq
Sequencing: MiSeq 300 cycle v2 nano
Read lengths: Read 1 = 76, i7 = 3, i5 = 8
Final pool nM: 4
Qubit final pool (ng/uL): 5.6
TapeStation size: 295
Calculated nM: 28.76
Loading pM: 4
%phiX: 25
cluster density: 332
Loaded 4 pM cDNA library (25% of 1st run)
Added 25% PhiX (but coming back as 46%)
Aligned = 46% (Read 1)
**Suggests PhiX is clustering more effectively than what else in in the tube…
BaseSpace report
“QC Passed”
%Q30 = 93%
% Cluster PF = 54%
Density = 332 K/mm^2 – seems low but tile image appears higher


A bit of a dose-response curve (two failures also plotted—see below).

I have included both biological (different colours) and technical replicates (similar colours).
To my rookie eyes, it seems that samples Q11 and Q14 have failed, though it is unclear to me whether that was at the RT, PCR, or sequencing step(s).
Based on what I have read in this Group, failure of individual libraries is not unexpected from FFPE samples.
The remaining question is whether anything else might be done (e.g., size purification, changes in PCR cy cles, etc.) to improve the Cluster PF%?
My guess is that the primer dimers living in the Tapestation peak at 165 bp is sufficiently abundant to cluster efficiently, take up space on the lane, but not sequencing, to account for the low Cluster PF%. This would be a problem if it interferes with deep sequencing of the “real” library on the NovaSeq, though I am unsure how much to be concerned as this is our first rodeo.
With appropriately amplified gratitude.
--