Hello Bill,Do you obtain the same readings when measuring from the device directly?Could you please capture a screenshot of the SARK Plots screen and from the device as well, if possible?Thank you,Melchor EA4FRB
El jue., 30 abr. 2020 a las 16:27, william gravett (<wgrav...@gmail.com>) escribió:
--I am trying to measure parameters on my 160M vertical. I tried the osl calibration and all seemed to go well....however, when I connected to antennaand used SARK plots and computer control on the device, I get two horizontal parallel lines which I know are incorrect. (measuring vswr and z)I tried measuring my GAP vertical and also my hex beam and got bogus indications again.I had similar indications some years ago and the tech at steppIR said sounds you need a osl cal, which I performed and all worked fine.I use the 160m preset frequencies on sark plots.Any suggestions much appreciated.Bill, W5PGZ
You received this message because you are subscribed to the Google Groups "sark110" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sar...@googlegroups.com.
To unsubscribe from this group and stop receiving emails from it, send an email to sark110+u...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/09b4e4fb-0867-4016-845a-1444e6106903%40googlegroups.com.
Please follow the troubleshooting instructions described at: http://www.sark110.com/resources/repairs/input-stage
This will allow to determine if there is a hardware issue in your unit.
If results are ok, try to repair the internal drive of the sark110 using the Windows “scan and repair drive” tool (see http://www.sark110.com/resources/repairs/formatting-the-internal-drive). It may happen that there is an error in the drive and the calibration files not read correctly.
It will also help if you make a copy of the files in the drive of your unit and send to me for analysis.
Thank you,
Melchor EA4FRB
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/CAJRaGB9n8Lwoq_vU-SXCivgbhQAHxyXpKNV8GqHZbHorHVUs0w%40mail.gmail.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/09b4e4fb-0867-4016-845a-1444e6106903%40googlegroups.com.
--
You received this message because you are subscribed to the Google Groups "sark110" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sar...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/CAJRaGB9n8Lwoq_vU-SXCivgbhQAHxyXpKNV8GqHZbHorHVUs0w%40mail.gmail.com.
I can provide a set, but the calibration files have coefficients adjusted for the specific unit and connector arrangement, so they would not work properly on your unit. Anyway, you can solve this problem by performing the Detector Calibration and then the OSL calibration; from your previous emails I understand that you have a OSL cal set. Please let me know if the problem is solved after calibration.
Melchor, EA4FRB
To unsubscribe from this group and stop receiving emails from it, send an email to sark110+u...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/bccc4e61-489e-4573-a395-b75ccb0a850a%40googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/008101d61fe9%24c8227230%2458675690%24%40gmail.com.
Could you send me the files of the internal drive of your device to check if the contents of the calibration files and settings are ok?
Thanks,
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/CAJRaGB-rQ%3DMQkucEx%3DdBkGgR6SFU9NOBiUGHWNVp90VfR3HT8Q%40mail.gmail.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/001101d620a9%2402b0ee80%240812cb80%24%40gmail.com.
Hello Bill,
Thank you for the files. I’ve observed that the OSL calibration file (oslcalib.dat) has wrong coefficients. I don’t know why, but it looks like the load used for the open condition has a short; at least, I was able to reproduce a similar effect using a short load as the open load during OSL cal. Is it possible that you have a defective open load? But in this case I wonder why the load indicator did not show a bad condition.
As an immediate workaround just for test, you could replace the calibration file by the attached one that I’ve prepared (this is calibrated in the same freq range from 1.8 to 2MHz). The readings will not be fully accurate, but at least the indications should not be stuck to VSWR:1.0 and |Z|:50.
Best regards,
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/CAJRaGB_2N7g%2BaKteu9_iJzu1ZMUF2A2m%3D437Kp9f7%3DmiOj7G%2Bg%40mail.gmail.com.
--
You received this message because you are subscribed to the Google Groups "sark110" group.
To unsubscribe from this group and stop receiving emails from it, send an email to sark110+u...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/sark110/de6724d5-93c4-4c7b-85fa-b1ca9eeeb8ee%40googlegroups.com.