Hi Lee
I recommend you do a baseline guiding to make sure you (and we) understand the basic performance of your mount. The guide to do this is attached
first you are calibrating very far from dec 0, so that is almost certainly causing issues
please download the latest DEV release and use the calibration wizard
your calibration also had a number of lost star events, so make sure you get a clean calibration before proceeding.
strainwave mounts generally are guided differently than GEMs, here are a few suggestions to start:
- fast exposure/guiding cadence of 0.5-1 second
- do not use PPEC for RA, instead use Hysteresis (at least for now)
- you may need to experiment with the aggressiveness of the RA algorithm.
I would avoid using 'vibration suppression pads', and make sure you are on solid concrete or other firm surface (i.e., not grass or soil). make sure you are not walking around near the mount.
that's a good start