When you get out again, make sure you accept all GA (guiding assistant) recommendations. It appears from my initial inspection that the backlash compensation, at a minimum, from the GA was not accepted. If GA didn't give you a checkbox for backlash compensation, look for potential issues which cause the apparent backlash measured by GA to be reduced to a value that does provide an accept button.
Hi Mine. I think the situation is as you suspected, the very poor mechanical performance of your Dec axis is causing your imaging problems. Here’s a typical sequence of guiding showing only the Dec in green:

What you see here is a long delay before the axis reverses direction followed by an immediate over-correction in the opposite direction. This often implies two problems: a large backlash plus stiction (static resistance) on the axis. We can’t really distinguish between them, the whole sequence might be caused by just binding and stiffness of the Dec axis. But you can see the very bad effects this will have on your imaging – each of these cycles is letting the guide star wander around by as much as 15 arc-sec. Unfortunately, many other users have encountered these sorts of problems with brand-new iOptron mounts. As a first step, perhaps you should go back to iOptron and ask them to replace the mount or help you fix it. If that doesn’t work out, you can probably do some reasonable imaging by guiding in only one direction for Dec. It looks like you might have tried that at one point but got the direction wrong. There’s a fairly length discussion in the Help guide for how to do uni-directional Dec guiding. But if it’s a new mount, I would start by asking the manufacturer to correct the problem.
Hope this helps,
Bruce
--
You received this message because you are subscribed to the Google Groups
"Open PHD Guiding" group.
To unsubscribe from this group and stop receiving emails from it, send an email
to open-phd-guidi...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/open-phd-guiding/eaeee1f9-8357-4d04-87a9-c1e822281a32%40googlegroups.com.
To unsubscribe from this group and stop receiving emails from it, send an email to open-phd...@googlegroups.com.
As you know, the response from Teleskop-Express is just wrong. The mount *does* respond (eventually) to guide commands in all 4 directions so there’s nothing wrong with the USB cable. And of course no errors are reported when PHD2 communicates with the mount through the night. Maybe they should stare at the graph I sent you and re-think their response.
The basic approach to uni-directional guiding goes like this:
If the mount is only a few months old, I think you should demand a refund from the manufacturer but that’s just me.
Good luck,
To unsubscribe from this group and stop receiving emails from it, send an email
to open-phd-guidi...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/open-phd-guiding/c1bd1703-00bf-466c-9c43-f3a75dff0470%40googlegroups.com.
On Apr 22, 2020, at 10:58 AM, bw_msgboard <bw_m...@earthlink.net> wrote:
As you know, the response from Teleskop-Express is just wrong. The mount *does* respond (eventually) to guide commands in all 4 directions so there’s nothing wrong with the USB cable. And of course no errors are reported when PHD2 communicates with the mount through the night. Maybe they should stare at the graph I sent you and re-think their response.The basic approach to uni-directional guiding goes like this:
- Don’t hyper-tune the polar alignment – alignment error in the 5-10 arc-min range will make things easier for you. It will also be easier if most of your polar alignment error is in altitude.
- After calibration, run the Guiding Assistant for a long enough time that you can clearly see the direction of drift *in Declination*. If the direction is “up” on the graph, you want to do *south* guiding and vice-versa.
- Expect that you will need to switch direction at some point in the sky. If most of your polar alignment error is in altitude, that will happen sometime around when you do a meridian flip.
- You can use either Resist-Switch or LowPass2 algorithms for Dec but don’t be overly aggressive with them. The guide star may never quite reach the starting point, but that doesn’t matter. What you care about is not getting reversals in direction and getting a Dec RMS value that’s reasonably close to RA. On most mounts, that isn’t hard to do.
If the mount is only a few months old, I think you should demand a refund from the manufacturer but that’s just me.Good luck,BruceFrom: 'Mine Be' via Open PHD Guiding [mailto:open-phd...@googlegroups.com]
Sent: Wednesday, April 22, 2020 9:37 AM
To: Open PHD Guiding
Subject: Re: [open-phd-guiding] Bad Performance in DEC while Guiding, iOptron CEM25P
Hey,thanks for your answer!
Isolating the Dec axis in the graph makes it look horrendous sadly...I'm from Germany, so i tried to talk to my vendor (teleskop-express.de) today where I ordered the mount, they suggested a broken USB/Serial cable which I don't believe as I did my fair share of research to order one that's supported under Windows 10 and is durable and high quality, some other amateur astronomers use the same one I ordered, I don't have any driver problems or random disconnects either. I tried to explain what I suspect but they waved off, saying the mount does track normally (which it does) so there can't be anything wrong. Sending to iOptron from Germanyisn't an option for me due to high shipping costs. I'm aware there's a PDF file from iOptron Technical services (https://drive.google.com/file/d/1-wSYganf13_zmSb7mHK5TUUqCZalP5n8/view) which shows how you can tighten the belt, I'm gonna check the belt tension now, and try to do it myself, fingers crossed.
To view this discussion on the web visit https://groups.google.com/d/msgid/open-phd-guiding/FB41BC1615504B278887E6AF8102681A%40HomeDesktop.
To view this discussion on the web visit https://groups.google.com/d/msgid/open-phd-guiding/c1bd1703-00bf-466c-9c43-f3a75dff0470%40googlegroups.com.
Hi Mine, this is really good news, nice job! Obviously there’s still a lot of reversal delay in the axis but the PHD2 Dec backlash compensation seems to be controlling that quite well. You can probably leave well enough alone at this point and just do some imaging. But one thing you could consider doing is to increase the guide speed setting in the mount from the current 0.5x sidereal to a value closer to 1x. That could further reduce the reversal delay by a proportional amount. If you do that, just be sure to then repeat both the calibration and the GA backlash test. In any case, the backlash compensation algorithm has a lot of work to do with your mount, so you may find that it takes a little time for the Dec guiding to stabilize after you do a slew. So you might want to keep an eye on it and not start an image until you can see there are no oscillations in Dec.
Thanks for reporting back!
Bruce
From: 'Mine Be'
via Open PHD Guiding [mailto:open-phd...@googlegroups.com]
Sent: Thursday, April 23, 2020
4:03 AM
To: Open PHD Guiding
Subject: Re: [open-phd-guiding]
Bad Performance in DEC while Guiding, iOptron CEM25P
Hey, thanks for your answer. I opened up my mount yesterday in the afternoon and tightened the belt in the Declination axis which was very loose. That fixed it for now, I'm down to 2700ms backlash according to the guiding assistant which could be compensated by turning on backlash compensation with the recommended values, I was hovering between 0.5-0.8" in both Axis last night, so finally the mount performs as it should. I went with resist switch for now, which worked great, turned the aggression down to 80 I think. Thank you for explaining unidirectional guiding to me, very helpful just in case it gets worse, but last night was really successful.
I couldn't return the mount as it's already 3 months old, and the vendor doesn't offer refunds in that case as he said the mount is working and it's likely the cable. That angered me quite a bit and I went with the belt tightening procedure.
All subs were great from last night, except the last 20 because a tree got in the telescopes way, but even then, guiding was still okay as the guiding cam apparently got a glimpse through the leaves, I'm really impressed.
Am Mittwoch, 22. April 2020 19:58:06 UTC+2 schrieb bw_msgboard:
As you know, the response from Teleskop-Express is just wrong. The mount *does* respond (eventually) to guide commands in all 4 directions so there’s nothing wrong with the USB cable. And of course no errors are reported when PHD2 communicates with the mount through the night. Maybe they should stare at the graph I sent you and re-think their response.
The basic approach to uni-directional guiding goes like this:
1. Don’t hyper-tune the polar alignment – alignment error in the 5-10 arc-min range will make things easier for you. It will also be easier if most of your polar alignment error is in altitude.
2. After calibration, run the Guiding Assistant for a long enough time that you can clearly see the direction of drift *in Declination*. If the direction is “up” on the graph, you want to do *south* guiding and vice-versa.
3. Expect that you will need to switch direction at some point in the sky. If most of your polar alignment error is in altitude, that will happen sometime around when you do a meridian flip.
4. You can use either Resist-Switch or LowPass2 algorithms for Dec but don’t be overly aggressive with them. The guide star may never quite reach the starting point, but that doesn’t matter. What you care about is not getting reversals in direction and getting a Dec RMS value that’s reasonably close to RA. On most mounts, that isn’t hard to do.