On the plate tests:
As all labs are supposed to calibrate their heaters using heat flux gauges, that are presumably more accurate ways of heat flux metering than the plate, the plate tests should not show any differences. In theory. In practice, of course, this would be a relatively easy way of checking the consistency of assumed incident heat fluxes. Doing this test before a flammability tests would be a verification test for the experimental work.
As mentioned already, the plate cannot reveal differences btw setups that are caused by more detailed characteristics of the incident radiation. For example, we currently investigate if the different source temperatures in cone and FPA, in the interaction with the PMMA absorption spectrum, can explain why cone and FPA give different burning rates. Plate verification could still claim that these tests were identical!
So, plate will be useful verification, but not sufficient for removing differences.
Simo
--
You received this message because you are subscribed to the Google Groups "MaCFP Discussions" group.
To unsubscribe from this group and stop receiving emails from it, send an email to
macfp-discussi...@googlegroups.com.
To view this discussion on the web visit
https://groups.google.com/d/msgid/macfp-discussions/20e575c0-525b-46b1-9675-10f5eed758cfn%40googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/macfp-discussions/78cabc7121be41b7b950f3a972ba36c3%40aalto.fi.
I tend to agree with Fabian. I think plate thermometers have their own issues. This discussion implies that cone should be our standard reference measurement, a benchmark for our models. Are we certain about that? Personally, I think that an inert environment radiation-driven gasification experiment represents a better benchmark for pyrolysis. However, I also realize that few labs have this capability.
In any case, if we decide that cone is our benchmark and we need to harmonize these measurements between various labs, I suggest that we start with something simple. Let us run 50 kW/m2 tests of our black PMMA and ask everyone to wrap the sample in aluminum foil, place it directly onto Kaowool PM insulation (at least 1 cm thick) and make sure that the sample is positioned 25 mm from the cone heater. No edge frames or any other contraptions should be used. We can distribute the insulation and foil to make sure that everyone uses the same materials. Let us also use the same standard duct flow rate of 24 L/s.
This approach will eliminate sample holder uncertainties and we will be left with uncertainties in oxygen consumption measurements and radiant heat flux. The uncertainties in the in oxygen consumption measurements will manifest themselves in the heats of combustion values obtained through integration of HRR. The heat of combustion should be between 24 and 26 kJ/g. If it is not, we know that the measurement is problematic and where the problem is. We can then re-normalize all HRR curves to produce the same heat of combustion, let us say 24.5 kJ/g. The remaining scatter in the average and peak HRR can now be attributed to uncertainties in the radiant heat flux settings. If all labs also send us the heater temperature data, which should be available in any standard cone, we should be able to figure out whether it is the calibration of the gauge or spatial non-uniformity of the heat flux that is responsible for each observed discrepancy.
Stas
The reason I have not considered the flat plate, is the question of how confident we are that it will keep its properties after several years of using it.
In the past I have used the time to ignition (or even time to Peak HRR) of PMMA, to verify the heat fluxes, but could not now do that since they changed its properties over the years. So, unto the Rexolite PS…
Anyways, I am quite satisfied with the pure ethylene glycol (EG) liquid, as it provides a nice way to check out the full functioning of the cone calorimeter in a single test (and also the FTIR, as it has EG in its characteristics, especially during the pre-ignition period). Since I am interested in wood and vegetation, the combustion water vapor emitted is also of interest, and EG is cooperative in the calibration for that, and has a behavior as a liquid that is most like the wood.
The weigh scale is also an important part of our cone calorimeter calibration, and with its fast response, the digital exponential smoothing needs to be done to its raw data to match the T90 responses of CO2, CO, and O2, and for that the EG combustion is cooperative in that it does not char, and hardly produces the black smoke. Our H2O sensor is not as fast responding as the CO2, CO, or O2 responses, so we use a digital deconvolution sharpening with the H2O sensor data. So if you have a pyrolysis process that is very fast, and model it with the mechanistic kinetics, then you will need to also account for the T90 responses of the cone sensors, for more accurate assessment of the pyrolysis and combustion modeling.
This question of radiation absorption differences between FPA and the cone calorimeter, would be interesting to explore with the EG.
Mark
|
From: macfp-di...@googlegroups.com <macfp-di...@googlegroups.com>
On Behalf Of Brännström, Fabian
Sent: Friday, April 30, 2021 3:10 AM
To: MaCFP Discussions <macfp-di...@googlegroups.com>
Subject: [External Email]Re: MaCFP-2 Condensed Phase Workshop - Developing requirements for data set quality
[External Email]
If this message comes from an unexpected sender or references a vague/unexpected topic;
Use caution before clicking links or opening attachments.
Please send any concerns or suspicious messages to:
Spam....@usda.gov
To view this discussion on the web visit https://groups.google.com/d/msgid/macfp-discussions/43850466-BB3F-4893-8354-F39C044345BD%40uni-wuppertal.de.