I don’t think the multi-band factor would interact with the likelihood of PNS. Something like the echo spacing is more likely. I get PNS pretty frequently in the scanner, but I am tall and broad shouldered.
In any case, we would recommend a somewhat different MRI protocol for the fMRI these days:
2.4mm isotropic
TR=1s, FA=60
60 slices
Optimized echo spacing (minimum or close to minimum)
This is all achievable with a MB=4 and the SNR is better than what we ran before without compromise for denoising or spatial localization.
Matt.
--
You received this message because you are subscribed to the Google Groups "HCP-Users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to
hcp-users+...@humanconnectome.org.
To view this discussion visit
https://groups.google.com/a/humanconnectome.org/d/msgid/hcp-users/de2191f7-d34e-46e8-8663-5d6c4ebd4fddn%40humanconnectome.org.
The materials in this message are private and may contain Protected Healthcare Information or other information of a sensitive nature. If you are not the intended recipient, be advised that any unauthorized use, disclosure, copying or the taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this email in error, please immediately notify the sender via telephone or return mail.
To view this discussion visit https://groups.google.com/a/humanconnectome.org/d/msgid/hcp-users/1A8115A1-AD0B-462F-B46C-ED9DE595CDD5%40wustl.edu.
> I don’t think the multi-band factor would interact with the likelihood of PNS
Yes, because it is not the RF excitation pulse that causes PNS but primarily the rapidly alternating EPI read-out which does not differ between multiplexed and conventional EPI.
Thus, multiband EPI more easily runs into SAR but not PNS issues.
Cheers,
Andreas