Greetings:
There is a technical note discussing TDR probe rod length and resolution at
http://www.campbellsci.com/documents/apnotes/tdrrod.pdf.
Josep's approach is an interesting one and compares well with the approach
presented in the note.
My regards,
Jim
Jim Bilskie, Ph.D.
Soil Physicist
Campbell Scientific, Inc.
On 7 Jul 2006 at 4:22, torrents@eel.upc.edu wrote:
>
> Dear Oscar,
>
> I am not sure if I'll anwser your question. I will try to be short:
>
> 1) Usually, conductivity of sandy soils is low (compared with lime or
> clay).
>
> 2) TDR pulse speed is v=c/sqrt(k) where c is the speed of light and k
> is epsilon relative of the mixture of soil (k=2-5) and water (k=80).
> It is an average (e.g. k=10). 3) Your system samples at a rate of M
> GSps (Giga Samples per second). 4) Aditionally risetime (tr) of your
> pulse lowers the effective number of samples you detect inside your
> soil sample.
>
> Example: k=10 => srt(k)=3 =>v=(3*10^8/3) m/s, assuming tr=0, M=1
> GSps=(10^9) Hz
> distance inside soil sample= v*t/2 (/2 because pulse goes and backs),
> only one measurement inside sample=>(3*10^8/3)/(10^9)/2=5 cm
>
> At least, you'll need 4 or 5 samples inside, then your minimum length
> should be about 20 to 25 cm. Life is not so easy (tr) but you have a
> rough number. Then, you should check M and tr in your system.
>
> Sincerely,
> Josep M. Torrents
>
> ps. A colleague told me about a software in internet that do these
> calculations, but I never checked.
>
>
>
> List Owner - BruceM wrote:
> > Posted on behalf of Oscar
> > =========================
> >
> > Good morning to everyone
> >
> > I have a question concerning TDR probes. I am about to buy a TDR
> > system in order to use it in a column experiment but my problem
> > concerns the lenght of the rods to be used. In the brochure of the
> > TDR system i will buy it is given only an upper limit for the soil
> > conductivity for which a specific lenght for the rods might be used.
> > Since my column experiment will be 20 cm diameter I was thinking to
> > use a 7.5cm lenght rods but I was wandering if these might be too
> > short in order to have accurate readings of water content. Is there
> > any experience you might report me in using so short rods, any
> > problem encountered, expecially in using them in sandy-silty soils.
> >
> > Thanks to all of you
> >
> > Regards
> >
> > Oscar Cainelli <cainelli@ing.unitn.it>
>
>