Hi Phil---
Thanks for Your reply.
Some nice guy send me schematics of the power supply board (B16) and board B15.
Was that You???
I checked the Board B15 and found 3 shorted tantal-capacitors, C551, 548, 550.
Changed these capacitors solved the problem.
But there are some more problem with the instrument. In the 3 lowest range 300, 100, 30 micro Volt, the pointer of the movement is not stable. So I will
continue foult-finding. The final problem seems to be in the board B11. This board includes both divider and amplifier. With combinations of these divider and amplifier gives the 15 ranges of +50dB to -90dB.
OK, thanks once again for Your reply. I will continue the foult-finding.