Charles wrote:
>
> It quit working this week and found the 2 A fuse blown. Not in stock, so
> had to order some. In the meantime, took it apart and powered it from a
> bench supply and found it OK but only if the bench supply was turned on
> after the front panel switch was turned on. Turning it on or off with the
> front panel switch activated the crowbar circuit.
>
** The HP counter has a conventional PSU delivering about 14VDC followed by a switching regulator for the 5V rail. 2A fuse is between the bridge rectifier and a pair of 2200uF electros. If the 5V rail goes out of tolerance, an SCR is triggered to short the electros and blow the fuse.
> I suspected a switcher problem and probing in circuit showed C36 (the main 5
> volt switcher cap) to be measuring about 1 uF and the schematic shows 390
> uF. Replaced it and all is fine now.
>
** The 390uF electro is the filter for the 5V rail and comes after a choke, high speed diode and switching transistor.
With a value like 1uF it could not possibly do its job and the 5V rail would be full of HF switching noise which could trigger the SCR.
Nice of HP to fit that crowbar and save all the digital circuits.
I have worked on expensve gear that had massive damage or was a complete write off because of a wayward 5V rail.
.... Phil