Tektronix 2467B (similar to 2465B): Error Message TEST 04 FAIL 13 (3453)

29 views
Skip to first unread message

tek_user

unread,
Dec 1, 2006, 12:36:58 PM12/1/06
to
Hi,

i am owning an Tektronix 2467B. Maybe someone can give me some advice,
please.
The 2467B shows following message:
TEST 04 FAIL 13 (3453)
I have got the complete service manual, but it says something like
"Option Error" ???
(The scope includes Option 10 and 11.)

If i delete the error message by pressing the A/B tigger button, the
device works, but
it has a awful DC-Offset on all channels (which varies with the V/div-
selection).


Best regards from Berlin

Harald

Jim Yanik

unread,
Dec 1, 2006, 9:22:54 PM12/1/06
to
"tek_user" <bulli...@gmx.net> wrote in
news:1164994618.8...@79g2000cws.googlegroups.com:

Did you try running a DC balance? (IIRC,hold down both AC/DC/GND buttons
until the DCBAL procedure starts)

IIRC,"test 04" failures involve NVRAM memory retention.(all your cal
constants)

You may have to replace the NVRAM and recal the entire instrument.

Or you have bad preamp hybrids.

--
Jim Yanik
jyanik
at
kua.net

tek_user

unread,
Dec 2, 2006, 10:11:39 AM12/2/06
to
Hi Jim,

thank you very much. After running the Dc-balance the offset is ok !
...but the error message is still the same.
I forgot to say, that after deleting the error message,
there are a lot of question marks on the display:
for example

50mV???5V????......

Is it possible to post a foto in this group ?
I would be very glad, to get some advices.
Sorry i was wrong it is an 2467 not 2467B.

Thanks

Jim Yanik schrieb:

tek_user

unread,
Dec 2, 2006, 2:05:58 PM12/2/06
to
Hi,

i forgot to say that after runnig the DC-balance cal. the error message
changed to :

TEST 04 FAIL 03 (3453)

The following things are still up to date:

. After running the Dc-balance the offset is ok !

> I forgot to say, that after deleting the error message,

Jim Yanik

unread,
Dec 2, 2006, 2:01:08 PM12/2/06
to
"tek_user" <bulli...@gmx.net> wrote in
news:1165072299....@16g2000cwy.googlegroups.com:

I'd say you have to perform a full recal,and if this occurs again,your
NVRAM IC is bad,needing replacement and another recal.

JW

unread,
Dec 4, 2006, 5:43:18 AM12/4/06
to
On 1 Dec 2006 09:33:51 -0800 "tek_user" <bulli...@gmx.net> wrote in
Message id: <1164994431.5...@73g2000cwn.googlegroups.com>:

>Hi,
>
>i am owning an Tektronix 2467B. Maybe someone can give me some advice,
>please.
>The 2467B shows following message:
>TEST 04 FAIL 13 (3453)
>I have got the complete service manual, but it says something like
>"Option Error" ???
>(The scope includes Option 10 and 11.)
>
>If i delete the error message by pressing the A/B tigger button, the
>device works, but
>it has a awful DC-Offset on all channels (which varies with the V/div-
>selection).
>
>
>Best regards from Berlin
>

You should ask this in sci.electronics.repair. There's a fellow who posts
there who used to work for Tektronix. (sci.electronics.repair added)

Andreas Tekman

unread,
Dec 11, 2006, 2:36:35 AM12/11/06
to

JW schrieb:

Well, we also read here :)

Looks like a problem in the calibration constants, stored in the NVRAM
A "TEST 04 FAIL 13 " condition means according to my manauls a "bad
checksum in the NVRAM".

If the NVRAM is ok, than this error can be corrected by doing the
calibration procedure (all steps mentioned in the handbook) and storing
the constants in the last step into the NVRAM.

hth, greets from Stuttgart,

Andreas

Andreas Tekman

unread,
Dec 11, 2006, 2:37:39 AM12/11/06
to

Andreas Tekman schrieb:

Ah , sorry, noticed the follow up too late.

Andreas

tek_user

unread,
Jan 8, 2007, 11:43:25 AM1/8/07
to
Hei Andreas,

kannst Du mir bitte Deine E-Mail-Adresse mitteilen ?
(Die hier im Forum hinterlegte funktioniert leider nicht.)

Gruss

Harald


Andreas Tekman schrieb:

Andreas Tekman

unread,
Jan 9, 2007, 2:14:33 AM1/9/07
to

tek_user schrieb:

> Hei Andreas,
>
> kannst Du mir bitte Deine E-Mail-Adresse mitteilen ?
> (Die hier im Forum hinterlegte funktioniert leider nicht.)
>
> Gruss
>
> Harald
>

Nicht leider, sondern das ist von mir so gewünscht und beabsichtigt.

GumpyGus

unread,
Mar 28, 2023, 1:45:07 PM3/28/23
to
It is likely to be a run-down battery. The calibration constants are held in a CMOS RAM which needs battery voltage to hold the calibration constants.

--
For full context, visit https://www.electrondepot.com/repair/tektronix-2467b-similar-to-2465b-error-message-test-04-fa-85746-.htm

Peter W.

unread,
Mar 28, 2023, 1:49:42 PM3/28/23
to
Useful, but this goes back to 2007.