I have no idea what the primary voltages should be (As it turns out, about
16VAC. Weird!)
BUT, suddenly I hear a metallic sounding "tick!" and the unit begins
running. (I quickly measure the primaries and note the voltages JIC.)
sounds a bit like a relay, but there's nothing like that in any of the
power supplies (there's a number of them) that I can see. Figuring an
intermittent somewhere I begin jiggling connectors, tapping around with a
screwdriver handle, etc. Nothing. Meter keeps working.
Cold blasts from freeze spray, Heat gun, etc. Nothing.
Turn it off for an hour and back on repeating what I did above. Still
working.
Turn it off for a day and back on. Wash, rinse, repeat. still working
perfectly.
Things like this seem to happen to me often. Is it just me, or do you have
faults like this that drive you nuts?
I don't suppose anyone might have any schematics for this meter, would
they? I think at some point Datron sold off to Wavetek then Fluke ended up
with it, as the user manual is available from Fluke.
Thanks.
Electrolytics will intermit like this. Change all that are in the power
supply section.
But 'lytics do not explain the 5.8 VAC with no load.
G²
Does this unit use an IEC type power receptable with some sort of voltage
changing/fuse module that might need to be reseated or checked?
lets see..
~ DC = 5.8*sqrt(2) = 5.8*1.414 = 8.2;
So, with the head room required for a 5 V reg 78xx styles..
8.2-2 = 6.2 on the average.. I put in more than the normal
head room requires just to be safe on the cal's. This gives you at least
1 volt to play with...
This is all fine if you have good working caps on the device that is
giving you the DC inside, mainly the rectifier..
So, putting all that in context, I'd still be checking for the caps in
the power supply! Test the DC voltage on the input of the REG>>>>
Jamie.
All you can do, is leave the damn thing alone as long as it is
working, ignore the hassles, forget them and go on.
But to keep Mister Justin Case mollified, do not totally forget what
you did so that you can pickup the testing gauntlet where ever you
previously left off.
Not the case here, problem lies in the AC somewhere. When the problem was
happening, the AC voltage on the toroids primary was at about 6 volts or
so, when it should have been 16.
Will check that, thanks. This meter is a bit of a rats nets of wiring,
typical of other British test equipment I've worked on (Wayne Kerr for
example). I hope that Fluke cleaned it up when they took it over...
It is interesting to note that some of the other 5 volt supplies were
working, as well as some of the +-15 volt ones. They may be derived from
the other toroid, though.
We had a coil in a reversing relay that would intermittently not work
and cause the mechanical drive train to push off in never never land,
thus damaging some very expensive rotor cones. It finally got bad enough
at one point where we were able to fine it.
It turns out that this "GE" reversing contact assembly used pushed on
spades in side to make it's wire termination to the coil instead of
soldering them!>. And you had to unassembled parts not intended to be
unassembled to find this. The mechanical vibration of the unit would
cause a connection issue and not switch on the coil for the reverse
operation when needed... The other controls would just keep cranking the
motor trying to find the correct location when it was actually pushing
it out of sight! And no, there were no limiting sensors in this machine ;)