Comment #6 on issue 65 by
RZellerT...@gmail.com: SALPA not properly
Using the SALPA parameters you describe (45 sample half width, 0 post-peg,
6 minimum peg, 3 pre peg, 6 min symmetric, right?), SALPA was able to run
correctly on one of our rigs (AD rig).
However, SALPA behaved as you described in your figure if electrical
stimulation input was turned off. Note that this only occurred if I
disabled the 'Stimulation Timing' input in the 'Aux. Input' page of the
Hardware Settings. The Electrical Stim 'real time data stream' does not
need to be activated (on the 'real time' page of Hardware Settings), and
the Electrical Stim data data stream does not need to be recorded (in the
Recording Stream Setup dialog).
SALPA gets a signal from the electrical stim recording set up, which it
uses as way to determine when to blank the signal, IE when the signal is
too distorted by stimulation artifact to recover. When the electrical stim
input is not turned on, SALPA doesn't get that signal and just performs
it's usual curve fitting.
Given that you aren't recording electrical stimulation, that seems like a
probably culprit.
Thanks a bunch and I'm sorry about the extreme delay on this.