Historians believe that in newspost <kkq6h5$ilf$
1...@dont-email.me> on Fri,
19 Apr 2013, DK <
d...@no.email.thankstospam.net> penned the following
literary masterpiece:
>Another opinion, perfectly
>reasonable, is that the kind of tubes one uses really matters.
>What tubes do you use? I used the ones that worked really well
>on 2720 model and I don't think that they are made of polypropylene,
>for example. PE/ABI claim that plasticware requirements for 9700
>and 2700 series are identical. My problem stems from the fact that
>this is most definitely not the case. (Or I am having a bad dream).
We use what used to be ABgene tubes, now part of Thermofisher.
>
>>104C is pretty standard for a heated lid.
>
>Really? I thought 100C is the standard. But I really don't have experience
>with that many cyclers. Good to know it's not a fundamental problem.
It's not going to really matter. Polypropylene melts a long way above
that so anything approx. will do.
>
>>DK how are you determining overshoot, measurement by thermocouple in a
>>tube or watching the temp. The displayed temp is block temp and not tube
>>temp which, will always lag behind.
>
>In the initial testing, I stuck a thermocouple wire betwen a block and a
>thin-walled tube jammed in place. *Looked* accurate enough and
>of course I am not about to buy $2000 calibration kit (which just might
>to be the same thing...)
The problem with the thermocouple is that it will be measuring block
temperature and not the in tube temperature. There will be at least a
0.5C difference. Just as difficult measuring in the tube because the
sheer mass of even a tiny thermocouple will pull heat out of the small
volume in the tube.
A multiplex PCR across all wells and run out on an agarose gel is pretty
good at testing consistency across a block. Overshoot or undershoot
isn't usually too much of an issue providing it is consistent across the
whole block. I don't want a PCR working in the middle of a block and
failing at all the edges!