On 10/30/11 8:09 PM, amdx wrote:
> Hey Morris, is there a reason you don't want to stick a thermocouple in
> the tube for your temp measurements as others have suggested? It would
> sure simplify this section, and, you could get lots of design help on
> sci.electronics.design to raise the thermocouple output voltage nearer
> 5v at 537*C.
I may be able to do that but can't design it that way without first
determining the heater parameters. If the heater wire length takes up
all 16" of the available ceramic core channels, then there won't be any
room for anything else.
So far, everyone has been trying to jump ahead to the temperature
measurement part of the problem, but first we need to design the heater
portion in order to know how much space (if any) will be available for
the thermocouple bead and its wire leads.
Since I don't *know* that there's enough space for the thermocouple, I
pretty much have to plan to make the heater wire do double duty.
There actually is an advantage to this in that it limits the number of
terminals needed at the end of the tube to just two, instead of the four
that would be needed if there were separate heater and thermocouple leads.
"Make everything as simple as possible, but not simpler."
> Also I suspect you have thought that if the temp sustains itself or
> starts to run away, your Ardiuno will sense this and start limiting
> hydrogen input and/or pump more water.
I'm not planning to water cool this thing at all. If the reaction is
able to sustain itself, then the next step (phase 2) will be to coax it
to self-sustain in the 400°C neighborhood needed by the parent project
into which I want to embed the reactor.
The phase 1 testing is to find the (pressure,temperature1) points, if
any, at which ignition occurs and the corresponding minimum
(pressure,temperature2) points where the reaction becomes
self-sustaining. During the initial phase 1 testing the 'safe'
temperature threshold will probably be set down to 250°C to give me time
to get over being terrified. :-)
If a runaway condition is detected (either by exceeding the preset
'safe' temperature threshold or by a significant increase in radiation),
then the shutdown routine will [1] kill heater power (even if it's
already off), [2] shut off the hydrogen supply, and then [3] open a
valve between the reaction chamber and a vacuum tank to remove the
hydrogen from the reactor. A full shutdown should take less than a
second and the vacuum valve will remain open until an explicit "close
valve" command is entered by the operator (me).