Assume that you retreive a CR image from a PACS and modify the pixel
data. If you finally want to store the modified CR image (DICOM file)
to the PACS, which Data Elements do you need to update expect the
pixel data?
/Mats
this strongly depends on the operation performed on the pixel data.
You should even consider changing the SOPClassUID to Secondary Capture
(SC) which has lots of implications which fields must or may not be
taken from the original image.
What at least should be changed except for the pixel data is:
(0x0008, 0x0070) Manufacturer
(0x0008, 0x1090) Manufacturer model name
(0x0008, 0x0008) ImageType (must not be "Primary")
(0x0008, 0x0032) Image Time
(0x0008, 0x0018) SOP Instance UID
I'm not sure about
(0x0008, 0x0022) Acquisition Date
(0x0008, 0x0032) Acquisition Time
This list may not be exhaustive, and I am not too much an expert in
these things but all the experts seem to be absorbed by the IHE
connect-a-thon, so you now at least have a starting point.
Hope I could help a bit...
andromeda
c97...@cs.umu.se (Mats) wrote in message news:<e56e6970.03032...@posting.google.com>...
I'm not sure what to advise about the existence of private elements
in the orignal data set. If you have modified the image, I don't know
if you want to pass private elements on blindly or remove them under
the assumption that they might no longer describe your data.
Steve Moore
moo...@mir.wustl.edu
Mallinckrodt Institute of Radiology
andro...@gmx.net (andromeda) wrote in message news:<df5cf7c4.03032...@posting.google.com>...
"If the pixel data of the derived Image is different from
the pixel data of the source images and this difference is
expected to affect professional interpretation of the image,
the Derived Image shall have a UID different than all the
source images." PS 3.3 C.7.6.1.1.2
It sounds like this is one of those cases when the SOP Instance
UID should be changed.
I would avoid using the Secondary Capture SOP Class for
the derived image if it is at all possible to use CR, since
you will loose support for CR-specific features.
Removing the private elements is a good idea.
Image Type value 1 should be changed from ORIGINAL to DERIVED
as suggested, and the Derivation Description and Source Image
Sequence be set appropriately.
Content Date and Time should probably remain the same,
Acquisition Date and Time should definitely remain the same,
and Instance Creation Date and Time should definitely change.
What to do with the equipment description is now spelled
out in detail in the Contributing Equipment Sequence
description in the SOP Common Module (PS 3.3 C.12.1.1.5)
which was added in CP 270; as Steve points out at the
very least the Series Instance UID needs to be changed.
david